Turn-On Checkout Procedure
The following checkout procedure describes the use of the
front panel controls and indicators illustrated in Figure 2 and
ensures that the supply is operational:
a. Push LINE button to ON.
b. Set RANGE push button to desired range.
c. Turn VOLTAGE control fully counter clockwise to ensure
that output decreases to 0 Vdc than fully clockwise to
ensure that output voltage increases to the maximum output
voltage.
d. While depressing CC SET push button, turn the CURRENT
control fully counter clockwise and than fully clockwise
to ensure that the current limit value can be set from
zero to maximum rated value.