ok, this sounds good. Why. Built OUT all digitizer and Y ch 1-4, Choose only place ch4, ONLY put the digitizer and the Y ch in there and calibrate. do first the default procedure in cal menue. Then, in the cal menue see if you have high values, not more than 32000, if so hou have to calibrate DC and HF gain. Basically wait for warming up the Y ch to see if you reach the zero line, switching from GND to AC to DC in the 2mV range.(the Y ch is getting hot, dont worry, the relais are providing heat as most of them are activated) Hit the zero line key before. If there is no signifant moving when switching you can start a recal hitting the three buttons getting the cal menue. the cal procedure should not need more than 3 minutes for 4 ch. if he sees great values it will loop and loop again. also, on the digitizer change the 2 coils and all the caps !!! WHY ? These leaky capacitors hits down the supply voltage and are destryoing the DC filter coils, even when they loooks good. (measure the coil value in ohm, not more than 1 OHM is ok) good speed. tom