Variable knobs are at cal-position; in normal position. Knobs just were is wrong position since I just placed them back after putting case together.
Channel 2 needs to be calibrated again (both 1mV/5mV attenuation as well as variable attenuation/balance), since line on channel 2 moves when GND-coupling is used. But I doubt channel 2 calibration issues would crosstalk so much on channel 1. That said, it must be calibrated.
Also, according to service manual, gain is calibrated before variable balance., but good point still.
Fixed knobs now to proper place.
I did test now with 1kHz and 50Hz, no remarkable difference in performance:
1kHz/10Vpp.
50Hz/10Vpp
Due CRT slow sweep needed for 50Hz (or my skills with analog 'scope...) and camera framerate picture is what it is.
Voltage amplitude makes no difference, tried as low as 40mVpp.
As for my function generator, It's hooked like this. Not best way due improper termination.
All of my 'scopes are 1M and function generator is 50 ohm impedances.