Scope / PSU grounding issue?

Status
Not open for further replies.

riccardo

Member
Hi,

I have been trying sort out some ringing in a PWM circuit, but I am having trouble believing what I see on my scope.

Basially, the load is a 5R wire wound resistor which has one side connected to the PSU, the other side is is pulsed to ground via a MOSFET. The control circuit is powered via a 12V regulator.

I've noticed that as I turn up the voltage input, the ringing disappears. (or so I thought). But if you look at the attached scope shots, the voltage of the output does not match what I am setting on the PSU. I think the scope is telling me fibs!

I'm measuring with the scope between the common ground, and the load. Anyone know what might be happening?
 

Attachments

  • 13V.jpg
    353.7 KB · Views: 187
  • 20V.jpg
    389.4 KB · Views: 197
  • 30V.jpg
    376.2 KB · Views: 182
hi,
The wire wound resistor will have some inductance, unless its a non inductive type.
 
I know. I am not asking about the source of the ringing. I want to know why my scope might be displaying what it is. It does not seem to be right.
 
Is it a digital scope?

If so, you could be exceeding the voltage range of the sampling A/D
 
Yes digital (OWON PDS6062T). Hmm, I think u may be on to something there. If I connect the probe to my PSU, it stops measuring after about 15V. 15V seems ridiculously low for a measurement limit.

What can I do to get around this?
 
I would expect the oscilloscope to operate over at least the full range of the display (±20V on the 5V scale). If not then you may need to use a 10:1 probe to increase the maximum displayed input voltage by 10 times (50V/division on the 5V scale for example).
 
Last edited:
Yes, I am surprised it does not cover the full display.

I changed it to 10:1 and I can see now that turning up the voltage does not magically make the ringing disappear as it appeared to before! A fast diode on the output does the trick though.

Thanks!
 