Let's take one of your rows for the 6 VAC measurement:
6 VAC, In AC (Pin for IN), 1.8 (units ?) 1.8 (Units?), 0.4 (Units?), 0.4 (Units?). 0.4 (Units?)
This is supposedly the voltage at the IN pin of the AD726, but I'm not sure if this is an AC or a DC voltage.
What I'd like to see is:
A: 6 VAC, In AC (Pin for IN), A VDC, A VDC, , B VDC, B VDC. B VDC
and
B: 6 VAC, In AC (Pin for IN), A VAC, A VAC, B VAC, B VAC. B VAC
A is a variable 1.8 in what you measured. B is 0.4 Just don't know if it is AC or DC volts.
and both lines (IN, OUT) for nothing connected to the meter. e.g 0V and with all of the ranges.
My thought is that the attenuator has to be before the IN of the Chip, so we should see something like 0.1 VAC, 1 VAC 0.001 AC as you change the range.
Do, you know if the range is 1999 or 2999 or 3999 counts for full scale. My guess is 1999 counts full scale.
Units are important. For all I know, your measuring the values in "potatoes".
Be prepared to figure out where the Input of the IC comes from or the output goes to, but we're not ready for that.
There is no question that the output is stuck, but the input is stuck too except for the lowest range. I'm just not sure if your measuring the input to the AD736 in AC volts or DC volts?
Post a link to the manual of the clamp meter too.