but remember many electret capsules are not good at low frequencies.
Relates to:
Two 10uF capacitors in series produce 5uF. Then the -3dB cutoff frequency into 5.1k is 6.3Hz.
Healthy heartbeats can be well below 1Hz, particularly in super fit people
Very different from audio frequencies!
That frequency response plot down to 20Hz is hardly typical. I've spent hours in the past trying to find wide response ones from major electronics suppliers.
The first two capsules I looked at, at random, on Mouser
frequency (f) 100 20,000 Hz
Frequency Range 50 ~ 16,000 (Hz)
So arguing or criticising 6.3Hz response without considering the exact mic response seems a bit silly.
And:
While the repeat rate may be below 1Hz, the frequency components of the pulse waveform are obviously at rather higher frequencies or would be inaudible.
It got me curious.. I found an online reference library of heart & lung recordings:
[Note - turn the volume down before trying any of these, especially if you have a sub - it just about bounced mine off the floor].
**broken link removed**
Using the "First heart sound" example:
**broken link removed**
Zoomed in to a single pulse (first half) the longest interval between major peaks is roughly 25mS, so the lowest audible component is somewhere around 40Hz. Any cutoff frequency reasonably below that should be fine.
On the high frequency side, looking at a spectral analysis, that recording is approximately 30db below peak at 400Hz & 60db below peak at 1000Hz but has content to 7KHz. The spectrum plot also puts the lowest component at 43Hz.