your entire circuiit needs to be inside a steel case
A highpass filter passes high frequencies and blocks low frequencies. It has no effect on high frequency noise.
Three RC sections connected together have a loss of -29dB at the calculated cutoff frequency. It is droopy so if the calculated frequency is 480Hz then 60Hz will not be reduced much but 4.8kHz and even 9.6kHz will also be reduced a little.
Maybe hum comes from your power supply.
I beleive and others should correct me if I'm wrong that aluminium will not shield against EMF ?
have you created a ground loop by mistake ? all ground wires must ground at the same place or you will create small voltage differences in the grould voltage and in ampliers this can cause a lot of trouble
Yes of course, two opamps. I was thinking about somebody else's opamp that had no gain.Just to be sure I'm clear, you're suggesting using two LT1115s, one as an amp and one as a filter?
Each RC cuts the response 6dB/octave and 480Hz is 3 octaves above 60Hz. So 3 RC sections with an opamp create a slope of 18dB/octave with lower frequencies cut the most.If a high pass filter were calculated for 480Hz (I'm not sure where that number came from), it should still affect all frequencies below it as well right?
If the cutoff frequency is 480Hz then frequencies above about 1kHz will not have phase shift.wouldn't the phase of the higher frequency signals be affected?
Yes. The cutoff frequency will also have its phase changed.would only the phase of the lower frequency signals be changed?
The phase change is 90 degrees for each section.I did some calculations off some stuff from wikipedia and it seemed like there'd be a phase change of about 90 degrees.
Solving noise problems in a measurement setup must first begin with locating the cause of the interference problem. Referring back to the block diagram in Figure 1, noise problems could be anything from the transducer to the data acquisition device itself. A process of trial and elimination could be used to identify the culprit.
The data acquisition device itself must first be verified by presenting it with a low-impedance source with no cabling and observing the measurement noise level. This can be done easily by short circuiting the high and low signals to the analog input ground with as short a wire as possible, preferably at the I/O connector of the data acquisition device. The noise levels observed in this trial will give you an idea of the best case that is possible with the given data acquisition device. If the noise levels measured are not reduced from those observed in the full setup (data acquisition device plus cabling plus signal sources), then the measurement system itself is responsible for the observed noise in the measurements. If the observed noise in the data acquisition device is not meeting its specifications, one of the other devices in the computer system may be responsible.
Try removing other boards from the system to see if the observed noise levels are reduced. Changing board location, that is, the slot into which the data acquisition board is plugged, is another alternative.
The placement of computer monitors could be suspect. For low-level signal measurements, it is best to keep the monitor as far from the signal cabling and the computer as possible. Setting the monitor on top of the computer is not desirable when acquiring or generating low-level signals.
Cabling from the signal conditioning and the environment under which the cabling is run to the acquisition device can be checked next if the acquisition device has been dismissed as the culprit. The signal conditioning unit or the signal source should be replaced by a low-impedance source, and the noise levels in the digitized data observed. The low-impedance source can be a direct short of the high and low signals to the analog input ground. This time, however, the short is located at the far end of the cable. If the observed noise levels are roughly the same as those with the actual signal source instead of the short in place, the cabling and/or the environment in which the cabling is run is the culprit. Cabling reorientation and increasing distance from the noise sources are possible solutions. If the noise source is not known, spectral analysis of the noise can identify the interference frequencies, which in turn can help locate the noise source. If the observed noise levels are smaller than those with the actual signal source in place, however, a resistor approximately equal to the output resistance of the source should be tried next in place of the short at the far end of the cable. This setup will show whether capacitive coupling in the cable due to high source impedance is the problem. If the observed noise levels from this last setup are smaller than those with the actual signal in place, cabling and the environment can be dismissed as the problem. In this case, the culprit is either the signal source itself or improper configuration of the data acquisition device for the source type.
You can enclose the wires in shrink wrap tubing and then run that through a hollow wire braid which is then grounded..
I really would like to assist you but up to this point you have failed to help us, help you. At this point we can only speculate on what you have going on, as you have not provided any sketches, diagrams of your system configuration. At this point I think you may be chasing your tail and building filters may not be the solution as the real problem has not been identified. As of yet, you have not mentioned how your cable shield is grounded, is it like in fig 18 of app note I provided? Too many unknowns so I am at a loss. Is your high gain amp at the far end of the sensor or at the sensor side of things? Maybe you might consider using a differential driver and receiver for the sensor data. The differential method is good for noise immunity. Anyways good luck.
No, the proper solution here if you have a long cable run is a battery powered single ended to differential amplifier with some gain directly at the sensor, driving shielded twisted pair, to a differential receiver on the far end. High pass it then if you want to. But just making a high pass filter to get rid of the 60Hz is a poor solution.