Mains transformers, DMMs, Continuity Tests and Confusion
I just had a problem with my HP Function Generator, it was blowing mains fuses.
So to try and find the problem, I opened the case and started to do some continuity and earth leakage tests on the mains input circuits.
I ended up very confused as the DMM (set to Ohms and Autorange) was flickering wildly and would just not settle down to steady reading.
I was beginning to think that there were problems with burn insulation in the input voltage selector switch in the function generator, but something in my mind told me that I had seen problems like this before.
So, SET THE DMM TO A FIXED RANGE... NOT AUTORANGE...
... end of problem!
Nice steady resistance readings, exactly as they should be.
So, what was the problem with the DMM?
As it changes resistance ranges, so it changes the current flowing through the circuit under test.
This is fine for resistors, but when there is a nice inductive mains transformer in circuit, as the test current changes there is a back EMF from the transformer.
The DMM sees the back EMF and changes ranges, and so it goes on continually changing ranges and displaying nonsense.
(The original problem with the Function Generator turned out to be the mains switch, it was making intermittent high resistance contact when closed.
As there was a spare pair of contacts on the switch, I just moved the wires for the bad contacts to the good spare ones. Job Done!)
JimB