Bottom rails are ground and VCC. I used top blue rail for negative voltage and top red rail as another ground to connect shields of input cables. 8K resistor sets the gain and other two goes to leg driving op amp. I noticed that I forgot to put the actual op amp there and connected it directly to my leg which explains why it did not work. I did not sleep for the past 26 hours so yeah...I may be blind as a bat but I do not see where you tied the top and bottom + rail(red) and
the top and bottom ground rails (black) together ? Note the top red rail seems to have no
components in it so comment here may be irrelevant. When you jumper the ground rails
together make sure you do not create a large loop by doing it at extreme ends of proto board.
Is your lab equipped with florescent lighting, consider turning that off when testing.
The 3 R's used to set G, why not just 1 ?
Regards, Dana.
I fixed the problems with the circuit and applied the stuff you mentioned. I get some kind of reading with acceptable off set. I am not sure if it is noise or something else though. I think I may get it right if I get the correct substance to rub on the electrodes.I may be blind as a bat but I do not see where you tied the top and bottom + rail(red) and
the top and bottom ground rails (black) together ? Note the top red rail seems to have no
components in it so comment here may be irrelevant. When you jumper the ground rails
together make sure you do not create a large loop by doing it at extreme ends of proto board.
Is your lab equipped with florescent lighting, consider turning that off when testing.
The 3 R's used to set G, why not just 1 ? Using that approach to get the desired G that cant
be done with a single R ?
Just for drill make sure your ground rail from extreme left contact to right has continuity.....
it is in fact one buss.
Regards, Dana.
I got some sticky electrodes and applied them as suggested on #1 circuit 2 while using single AD620 at 1000 gain. I was finally able to get a reading on those. Using default electrodes on arms still cause the AD620 to saturate. It seems that getting the signal from arms is much more difficult since it is weaker and DC off set is apperent.I may be blind as a bat but I do not see where you tied the top and bottom + rail(red) and
the top and bottom ground rails (black) together ? Note the top red rail seems to have no
components in it so comment here may be irrelevant. When you jumper the ground rails
together make sure you do not create a large loop by doing it at extreme ends of proto board.
Is your lab equipped with florescent lighting, consider turning that off when testing.
The 3 R's used to set G, why not just 1 ? Using that approach to get the desired G that cant
be done with a single R ?
Just for drill make sure your ground rail from extreme left contact to right has continuity.....
it is in fact one buss.
Regards, Dana.
Fantastic waveform!I was finally able to get a reading on those.
And about 51bpm heart rateFantastic waveform!
I made the mistake of AC coupling it at low frequency AC signals before and seeing distortion. So it was in DC mode which is indicated on the display. If there is DC bias it shows up there which makes me believe it was the correct mode. Ground mode says GND on display.Your scope shot picture and its measured Vmin and Vmax dont seem to align
with the trace and what I think is the ground reference symbol in the display.....
You AC coupled ? Note if so AC coupled cutoff freq is usually in the 10 - 100 Hz
for many scopes, so take that into account.
PicoScope Oscilloscope AC Coupling
What is the difference between AC and DC Coupling, what is AC coupling used for and how does it work. Also brief test using DC power supply ripple.www.picotech.com
Open-DSO-150/documents/dso-150_Schematic_Shell.pdf at master · michar71/Open-DSO-150
An open source firmware for JYE Tech DSO-150 Digital Storage Oscilloscope - Open-DSO-150/documents/dso-150_Schematic_Shell.pdf at master · michar71/Open-DSO-150github.com
View attachment 139928
Regards, Dana.
I actually thought about just taping a piezo crystal over myself since intrumentation amplifier gain is insane it would %100 work if I decreased gain a bit. Two sticky electrodes worked fine and prof. said I can use them so I will probably ditch the arm electrodes. I read the same waveform for short periods of time from my arms but amplitude was greatly reduced and it immediately saturated the op amps after the salty water or cream evaporated from the electrode surface and my arms. It is extremely unstable with those clamp electrodes I think it is not worth trying to use them.I'm just now getting to this thread. My 1st thought regarding the electrode connectivity issue was to suggest using the sticky TENS unit electrodes. (I had one from physical therapy following an accident.) Replacement electrodes are available at most drug stores and disposable.
2nd thought: if push comes to shove, how about using a contact microphone taped to your chest? Trying to read & register the electro chemical microvoltages responsible for triggering the heart muscles isn't easy considering your source signal is filtered thru dry skin, wet muscle, bone & flesh. Who is to say you couldn't get "interference" from the muscles in between? Try flexing your pecks while it's attached and see if you done pick up false readings.
Just my 2 cents.
I am at least using shielded cables to probes. I honestly don't need it perfect, prof only wants a demonstration of that it works and it does enough. I would definitely solder it normally.The rows of contacts and long wires all over the place on a solderless breadboard guarantees lots of interference and oscillation.
Using shielded audio cables for the probes blocks interference picked up by them (antennas).
I can't really increase the gain of the AD620 since it also amplifies the DC off set of the electrodes and saturates with more than 10 gain. But I can get 100 gain at second stage and something like 2 at the last. Would that be more appropriate.The age old rule take your G in the first stage is a legit rule for minimizing
noise (this is not your exact application but discussion is appropriate) :
Page 8 specific recommendations.
Also with IA have best Voffset compared to your OpAmps then not Ging up
that Voffset with lots of G in that stage makes more sense.
Regards, Dana.
Well, the usual response from AG.The rows of contacts and long wires all over the place on a solderless breadboard guarantees lots of interference and oscillation.