Pressure Sensor for water level measurement
premkumar9
Hello,
I wanted to measure water level up to 50 Mts with 5 cm accuracy. Can I use a pressure sensor for this purpose? I tried a pressure sensor with built in amplifier got from one supplier. I kept it at a depth of 2.5Mts of water and monitored the output throughout day time. I found a variation of 1.5% in the output voltage. What could be the reasons for such variations? Could it be due to the atmospheric pressure variation? Is there any way to compensate it?
Your help will be highly appreciated.
Is the water in an tank with known size? Have you considered using ultrasonic technology to measure water level where tiny echos are transmitted and received?
Hello,
There are many ultra sonic sensor are not good in this situation, i think because it sensor get wet due to evaporation.
may be something like Electro magnet Reed switch?
Have you checked the specs of available pressure sensors? You are asking for a resolution of 0.1%. I'd be surprised if any one sensor type would be linear to that accuracy, so multiple calibration points would probably be needed. This link describes various non-linearity components.
Yes it is usually an expensive pressure transmitter that will give you 0.1% accuracy.
Can you tell us which pressure transmitter you are using?
Whether the Ptx will respond to atmospheric pressure depends on the type and configuration. More information required.
Hello,
While you do not mention the manufacturer or model number of the pressure sensor you used there are many variables that can effect your measurement over time. Temperature? Changes in atmospheric pressure? The uncertainty of the instrument used to measure the pressure sensor output? The excitation voltage applied to the sensor? Then you have the data sheet specifications of the sensor itself.
You can have the 0.1% accuracy you desire but it requires, as was mentioned, more expensive pressure transmitters. **broken link removed****broken link removed**. Just the transducer will cost about $2,300 USD and if we add a quality process meter we easily will exceed $3,000 USD. That is just an example, there are other manufacturers like Rosemount who make similar units in precision and cost.
So what this comes down to is I can buy a pressure transducer/transmitter for about $25 USD or I can buy a pressure transducer/transmitter for about $2,500 USD depending on my needs. They will both measure pressure but one does a much, much better job but cost much, much more. :)
What you are seeing is not at all unusual. I would start with a differential type sensor and work from there based on your uncertainty needs.
Ron
There should be some reference point on the top or face of the dam that doesn't move. Can't you measure referenced to that?
John
Your stated accuracy requirements are severe.
Given the depth of the water, you need a pressure sensor in the ≈70 PSI range (0.445lb per column-foot).
This site: http://www.astsensors.com/submersible_level_sensors/AST4500__AST4510
has sensors that, I believe, meet your needs (accuracy, barometric pressure compensation, etc), but are very expensive, especially with their associated electronics.
If I may ask, why the need for such accuracy (a 5cm depth differential would be overcome with a stiff wind "piling" up surface water)? jpanhalt's suggestion sounded inexpensive and easily implemented, if you're willing to accept a reduced accuracy.
Thank you all for your replies. Normally Alerts come to my inbox when somebody replies. This time alerts came only for first 2 replies. So I didn't check the other replies. Now only the replies came to my notice. I will go through them in detail and revert. Once again thank you all.
Many depth pressure sensors have a built in air vent pipe within the cable to the transducer.
This ensures that the differential between the water pressure side of the sensor diaphragm and the atmospheric side compensates for changes in barometric pressure.
The 2.5Bar depth transducer [~25mtrs] I used cost around ~£120, you can get 5Bar sensors [~50mtrs], which is the depth that you have specified.
The 2.5Bar transducer gave water depth accuracy of +/-5cms.
What maximum range of water depth 'change' do you expect to see in the dam water.???
E
For the original question; yes, a pressure transducer can be used. But I think that attaching a measuring tape to a dam wall would be much simpler, reliable, accurate, precise and long lasting.
The data is to be sent to the server. So electronic measurement is required.
I can fix an another pressure sensor at fixed location near the dam and take the readings. Can I use this readings to compensate the pressure variation due to atmospheric pressure change? Is this atmospheric pressure change the main reason for the drift in sensor output?
You could use an external atmospheric pressure sensor to correct for changes in the readings, but the 'drift' may be due to other causes in the equipment being used
It should be easy enough to test using the atmospheric pressure sensor,, record both the readings from the submerged sensor and the atmospheric sensor over a few hours,
observing the actual water 'height' using a semi immersed calibrated water level board.
Do you follow that OK.?
E
The datasheet Eric linked to had, pressure, density @ temperature and G.
I THINK if one were to look at relative height, you would need a ported sensor or one related to differential pressure. So, i would think you would need calm waters.
Acceleration due to gravity varies. See: http://www.blazelabs.com/f-u-massvariation.asp. Density will vary with temperature. If then sensor has any smart, the temperature would be measured at the sensor. The temperature may not be the same over the measured distance.
Further remember that the pressure REPORTED by the weather bureau is NORMALIZED to sea level even if your 5000 ft up.
You have a lot of potential error sources because this isn't a tank on the ground.
Now, I'm wondering if you could do a differential sort of measurement to a known depth (floating sensor).
Say the floating sensor is 1 foot under water, but it reports a value of 13". Your sensor reports 26' The density of the water may change with temperature and depth.
So, you sort of normalize it.
These are just ideas.I haven't worked in the field. I have had hydraulics and relativity in class, but don't remember much.
You may get a better understanding by reading this PDF.
E
  • wardwell.pdf
I understand like this. The actual pressure sensor output from the sensor immersed in water at a known depth and the output voltage from a sensor at fixed position has to be compared over a period of time to understand their relationship. From this a correction factor can be worked out. Am I right?